21.3% O2 reading not changing

Discuss 21.3% O2 reading not changing in the Gas Engineers Forum area at PlumbersForums.net

Messages
37
My testo 300 today when commissioning a new install would only give a reading of 21.3% o2 for flue integrity, tried zeroing outside but stayed at 21.3 tried turning on and off still the same. When in the flue gas test port it would go down to what you would expect and 5.3. Anyone have any ideas why a high o2 reading? Was only calibrated in feb
 
Seems the O2 cell has failed has it been used on a bad boiler with high co /co2 ? Hard to say otherwise I have a 327 and it has happened on mine a long time ago now .
 